MedWrench is available to help people with specific questions and troubleshooting problems. Could you please be more specific with the issue that you are having with your instrument? The MedWrench team will do their best to assist you in solving the issue. This will also help other community members who face similar challenges.
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
Richard Wolf 2075U Test procedure.
1. Connect black test jig to bipolar generator.
2. Connect black banana plug to dispersive blue connection on ESU Analyzer.
3. Connect white banana plug to active yellow connection on ESU Analyzer.
4. Connect green ground banana plug from ESU Analyzer to alligator clip ground point on Generator.
5. Turn on ESU Analyzer and select Manual--Output then select 100 Ohms as the test load.
6. Turn on Generator and use HF- Power knob to number 1.
7. Step on footswitch. Reading from generator should be 1.9 W +/-5%.
8. Select 2. Step on footswitch. Reading from generator should be 4.4 W +/-10%.
9. Select 3 on HF-Power knob. Step on footswitch. Reading from generator should be 9.3 W +/-10%.
10. Select 4 on HF-Power knob. Step on footswitch. Reading from generator should be 17.6 W +/-10%.
11. Select 5 on HF-Power knob. Step on footswitch. Reading from generator should be 29.7 W +/-10%.
12. Select 6 on HF-Power knob. Step on footswitch. Reading from generator should be 47.6 W +/-5%.