Bipap-Vision - PM testing DCS: F04 error code

I've replaced so much on this one unit already. Every test in the manual has passed, except when i look for error codes, I get a DCS:F04 error code, the manual says replace the ICB cable, PC Board, MC board, DCS Board. I'm just swapping out used parts here, has anyone experienced this error code?

   RE: Ventilator Report Post
Fri Apr 20 2018Reply from Chris-CBET
avatar placemark

follow up for future tech's needs.

 turned out to be the display board. not sure what specifically as everything appeared to work. 


Report Post
Wed Dec 30 2020Reply from Denova Medical
Denova Medical

Everything probably did work. I had the same problem and wasted time swapping boards. The software version (13.X) on the DC,MC and PC board need to match. If you swap boards and the software does not match you will get that error. You can exchange the EEPROM and the error will go away.


Photios Peter Dalamagas Denova Medical Inc (704)-233-7574