Tue May 01 2018
I need software for a Siemens Compact L
I need software for a siemens compact L
Tue May 01 2018Reply from Tomelgy
There is no an easy answer to that question because there are about 4 or 5 different versions with different levels of software which are NOT compatable with each other.
To help answer can you state which version of software required, do you want software for the Memoskop, for the Generator (main system) and so-on, SSW for Generator SSW for Memoskop??
The Memoskop has 5 different versions of software alone, it also depends upon which memosko, meaning Memoskop C, CE, C-Mod, C-Sub, CX and so-on.
The TV system was originall 512 X 512 matrix, later 1K X 1K....... 512 is not compatable with 1K.
To help identity your system and software it will be necessary to list the main system model (partnumber) & serial number, for extra clarity alos incliude the memoskop Model & serial number.
If you have the system manual (blue Binder from Siemens, it's all listed in there!
Tom
Mon May 21 2018Reply from ziad
Dear sir
send me ur e mail
Dear All any one can send me the compect L service manual and error code for the same siemens c Arm
Wed Apr 08 2020Reply from Irakli_khabe
Dear Sirs
Hello. It seems I have same problem with siemens siremobil compact. It is a problem of memoskop software. Can you Help to fix this problem?
Also the error number is 2001
Wed Apr 08 2020Reply from outinleft
Error E2001 System reboot after watchdog timeout
Possible trouble: Host registered a system reboot,
The watchdog on D1 triggered the reboot (system overload).
Possible help: Faults in the system.
Check power supply (voltage and ripple).
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Mon Apr 13 2020Reply from Irakli_khabe
Hello.
 here was problem in X10 connector's wires connection. We have repaired wires connection and System have started well after self reboot.
        Thank you