I have a cardiocap5 that I just replaced the NIBP cuff, hose and fittings. It will cycle up to 250 but then stops and comes back air leak. We have tried reconnecting everything but cannot clear it. Any suggestions?
Wed May 02 2018
NIBP error
Wed May 02 2018Reply from Alex Wright

Try checking the manuals listed here! I found this:
Problem: Air leakage Hose and/or cuff leaking. Damaged cuff, cuff connector, or hose double connector. Action: Replace cuff, cuff connector, and/or hose double connector.
Problem: Hose and cuff OK, but tube disconnected or damaged.
Action: Connect or replace tube.
Problem: O-ring damaged or missing.
Action: Replace O-ring.
Problem: Leakage inside the module.
Action: Replace the whole tubing.
Problem: Tube disconnected or damaged; valve(s) damaged.
Action: Fix connections. Replace tubes/valve(s).
Problem: Air chamber leaking.
Action: Replace the NIBP board.
If that doesn't help, check here.