INFORMACION ERROR 5035
Mon May 07 2018Reply from fundeehelectromedicina
Mon May 07 2018Reply from outinleft

Probbaly a bad Xray tube, or bad D2. We have both if you need parts. wayne@columbiaimaging.com
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Mon May 07 2018Reply from outinleft

Error E5035 Gen. IH actual value too low (standby)
Possible trouble: Actual filament current (IH_ist) in standby mode too low.
Possible help: Measure IH_ist at D1.
Check actual value cables.
Measured value acquisition on D1 defective, possibly D2 defective.
Filament broken.
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com