Need the calibration kit but do not have enough machines to justitify the cost. I have the calibration gas and regulator but there is a flow meter as well. Does anyone out there have this kit that can tell me the specs on the flow meter
EtcO2 Calibration

It takes two gasses to actually calibrate the unit, a 5% and a 10%. I used the 5% from a cal gas for another monitor and a 10% which I happened to have for a Mindray unit. The flowmeter need to read to about 300 ml max. Be aware that the proceedure in the service manual requires a good read thru. I found that one first has to check the cal, thru a specific proceedure then perform the cal, again thru a specific proceedure, and then check the cal AGAIN following calibration followed by a complete "OPERATIONAL CHECK" in order to clear the "calibration required" message. This same type of routine is required when calibrating the NIBP, as well. Good Luck. Note: manuals located in the resources tab of the product details at MedWrench.