Tube warm up stop when the OLP reach 62% tube heat temp go down

Hi sir, please help i have a problem of Toshiba asteion super 4 slice ct-scanner os windows xp platform TSX-021B when i power on the system host computer boot-up the Logo Asteion Blue bar graph finished it's 8 minutes wait before open the user main menu. second problem when i run Tube warm-up at the middle process stop exposure when the OLP reach 62% and tube heat temp going down fast gantry continue rotation few minutes stop and the computer hang - up but no error in the screen, sometime i run head scan procedure wait for the start button light not start What is service password?

   RE: CT Scanner Report Post
Tue May 15 2018Reply from Alex Wright
Alex Wright

Hello,

These companies may be able to assist you:

IMES

TriImaging 

RSTI



Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com

Report Post
Tue May 15 2018Reply from RichardsonHealthcare
avatar placemark

We are happy to help here. This should get you down the road. If you need more or help with service pws, be sure to call our 24/7 tech support line for anything Toshiba CT related! (704) 739-3597 

The problem is either a defective MHR board in the Rec. box or corrupt boot line settings on the RTM board in the Rec. box. Most likely a bad MHR board is causing the problem.

If a MHR board is defective (Missing from the configuration) the indications are as follow

  1. Boot up takes much longer than normal
  2. The Gantry/Couch icon and the tube OLP icon in the user interface takes a long time to appear in the user interface
  3. Start an exam
  4. The gantry will spin up to speed
  5. The Abort button on the keyboard will flash once, then the Confirm button in the user interface becomes active
  6. Click the Confirm button. The Abort button will flash once, then the Confirm button becomes active again
  7. The Start button on the keyboard never becomes active

You will be stuck in this loop until you quit the Exam.

Once you quit the exam, after about 10 minutes a message will appear on the screen. It will say "Error in pre-process (a bad parameter was sent to pre task)"

----Richardson Healthcare and IMES


Report Post
Sun Feb 27 2022Reply from Asteion 4
avatar placemark

Hi, I have same problem can you please help me 


Report Post
Mon Feb 28 2022Reply from RichardsonHealthcare
avatar placemark

We are happy to help. Were you able to follow the instructions in the thread? If you need further assistance, call our 24/7 tech support line (704-739-3597) or email us at healthcare@rell.com

-Richardson Healthcare


Report Post
Mon Mar 28 2022Reply from Mike Kiggins
avatar placemark

Sounds to me like you have a failing XC board. If the OLP readings are different or counting incorrectly it's going to be the XC. The XC will also stop from shooting XRays if it doesn't sense the correct return signals from the SRU-H and SRU-L. Same goes for the Spellman units. It would need a much closer examination to determine the exact issue. If it's intermittent now. It will go hard down soon.


Report Post