Siremobil Compact L does not radiate
Hello, I have a system that initializes normally but does not emit radiation and also does not give any error, I also can not work with the keys on the lower left side.
Thank You
RE: Mobile C-Arm Report Post
Fri Jun 01 2018Reply from Tomelgy

This sounds a little odd. If the radiation selection keys are not lit & no errors, this could be a Memoskop or communication between Memoskop & main Generator.
Check the wires in SG Cable (umbilical cord) for any problems would be the first place to start.
Tom
Report Post
Sat Jun 02 2018Reply from JorgePereira
Sat Jun 02 2018Reply from Tomelgy

Ok this is good. Next can you look at the generator D1 PCB Does the 7 segment display rotate clockwise or anticlockwise? If yes the D1 PCB has initialized OK. If no then you have maybe software or parameter file problem, Or power supplies, on the D1 PCB are test point lugs for +/-15v & 5v etc.
Check connectors are good, particularly the top ribbon cable which supplies the control console.
Tom
Report Post
Sat Jun 02 2018Reply from JorgePereira
Sat Jun 23 2018Reply from joedemormon

Please I have similar problem, how did you resolve yours? Siremobil compact l not emitting radiation, left lower keys not working. Any help will be appreciated.