Hi, Looking for a service manual for the Varic. The yellow rotational lock needs attention, won't keep the C arm locked.
Any advice appreciated.
Thanks. D
Thu Jun 07 2018
Rotational Lock
Thu Jun 07 2018Reply from Alex Wright
Check the operating manual listed here.
Fri Jun 08 2018Reply from Med Eng
HI,
Thanks for that, however its the SERVICE MANUAL that is required, not the operators manual which we already have.
There was another post for the Varic, requesting a service manual, however can't find it for the list of docs for the Arcadis Varic.
Was this ever uploaded.
Thanks D
Fri Jun 08 2018Reply from BMET-KY
There should be a small rubber cap adjacent to the locking handle on top of the column. under that you can adjust an allen to increase or decrease tension to lock.
Fri Jun 08 2018Reply from Tomelgy
A contributer explained to adjust the Allen screw under the rubber cover at end of horizontal arm, this is correct.
The factory recommended to use some locktite as the screw can come loose and in some cases fall out.
Tom