Alaris 8015 fails Fast Battery Conditioning
RE: Infusion Pump Report Post

Report Post

Report Post

the power supply could be the problem. However, it can be either the main supply or the switching supply-I have had this be the case for either supply
Report Post

There is an actual white paper on this very topic as to calibrating the batteries. The batteries should be replaced every 3 to 4 years as I am sure you already know, but the actual method to calibrate the batteries as seen on the recall on the issue is to run the unit with four pump LVPs all running at a fast clip so that the battery is fully discharged which will take roughly 28 hours, followed by fully charging the batteries to capacity for another 24 hours max. I can not post the recall paperwork on this here, but you can get a copy from Carefusion (BD) by calling them. This is for the 8015 and 8010 controllers. Hope that helps. We just replace the batteries every other year to keep from having our fleet running for days to condition the batteries as outlined in the recall notice..
Stan
mrstan.
Report Post

Check transistor Q19 on the power conditioning board, it is by the BATT_RAW connector. I found one that shorted out and caused erroneous hi battery voltage readings and prevented completion of the Fast Battery Conditioning routine.
Report Post