Although it is your choice to go offline with your discussion, we are sure that other MedWrench users would appreciate if you kept your discussion within the thread. This will help to keep a permanent record of the issue you are facing and how it was solved to help others in the community that may be facing a similar challenge.
We appreciate your participation and hope that the MedWrench forums will help you to solve your medical equipment issue.
Best regards,
The MedWrench Team
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
So what’s the status of this issue.
Jamison Ashby, MBA
Owner: Precision Imaging solutions
www.PI.solutions
Report Post
M326: what about the vcharger lamp, above the SAX1 connector on the stand? is the light very low?
if yes: charger problem.
you can try to reset the charger by:
switch off the system, disconnect the mains plug, disconnect the battery "-"pole, sometimes after 10s the chargerlamp will flash once. Wait 5 minutes, reconnect the battery, switch on the system. In many cases the system will work normal again. if not: exchange the ESU.
good luck!
Henk1951
Report Post
M326:
Henk, Thanks. Followed your instruction twice. Same M326 error message repeated. My vcharger lamp is completely off. When connect with SAX1 connector, it only blink once. Replaced the battery in early 2015. This not a battery issue.
Report Post
check wiring to connection area SAx1.
Just because you changed batteries 3 plus uears ago does not mean they are still good. Dont assume.
when you changed the batteries did you change the charger baord also? Reason I ask is that when you have tanked batteries you are working the cra_p out of the charger.
M Chang..are you in bellevue??
Report Post
are you still struggling with this problem? 326 can be a charger problem.
Do you have a logfile? You need BVScope + IST-key for downloading the logfile from the system.
Henk1951