Please can someone help me with the service manual of Siemens Mobilett II. Mine is displaying some error codes (12, 13 & 17) and I want to know the cause. Please you can send the manual to rayokonkwo@gmail.com. Thanks
Thu Jun 28 2018
Service Manual
Mon Jul 09 2018Reply from Alex Wright

Although it is your choice to go offline with your discussion, we are sure that other MedWrench users would appreciate if you kept your discussion within the thread. This will help to keep a permanent record of the issue you are facing and how it was solved to help others in the community that may be facing a similar challenge.
We appreciate your participation and hope that the MedWrench forums will help you to solve your medical equipment issue.
Best regards,
The MedWrench Team
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
Mon Jul 09 2018Reply from Chris Brown

Here is a breakdown of your errors. ERR 12 is most likely the 3.3V CMOS battery on the control panel. It will give the error when it goes below 2.5V. ERR 13 usually happens when they unplug the system without powering off. It sometimes causes the U2 fuse on the D6 board to blow. ERR 17 is the capacitors charging to slow. First make sure someone hasn't turned on the discharge switch. It is on the D7 board at the top front of unit. If that is fine then check big 101 fuse. After that check the fuses between the capacitors, but be careful of the 300VDC. If you get this far please make sure you discharge everything first. I use a bulb with jumper wires to discharge capacitor banks and measure everything before replacing bad fuses or bad capacitors.