Unit will work with AA batteries installed but it won't charge with a brand new charger. The charge light doesn't come on. Wondering if anyone has the service manual for this, all I can find is the Users manual. Does anyone know what the 2 components in the battery chamber are?
Thanks
Mon Jul 09 2018
Unit won't charge
Mon Jul 09 2018Reply from Alex Wright

You can also always contact Fluke Biomedical directly for more information or assistance.
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
Mon Jul 09 2018Reply from John S.

Hi,
Here are the possible solutions:
1) The four batteries must be replaced with our item number: 4481150, Fluke description: BATTERY ,SECONDARY, NI-MH, 1.2V, 2.05AH, SINGLE CELL, AA, BUTTON TOP. (Alkaline AA batteries will not charge.)
2) The proper battery charger must also be used. Fluke item number: 4766561, description: POWER SUPPLY, SW, 18W, 85-264VAC, 9V@2A, 3KV, VI, W/MAINS ADAPTERS, UNIV WALL MOUNT must be used.
3) If the above items are being used, then the IDA-1S and battery charger should be returned to the Fluke Biomedical Service Center. The number to call, for a Return Authorization is 440-498-2564.
Thank you.
John Shirlock
Fluke Biomedical Technical Support