Wed Jul 18 2018

Alarme 5

How can I remove this alarme please Al 5

   RE: Tissue Processor Report Post
Wed Jul 18 2018Reply from Alex Wright
Alex Wright

Hello, try checking this manual posted here. I think that information on alarms is posted on page 24 and after.



Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com

Report Post
Sun Jul 29 2018Reply from F. Maia
avatar placemark

Hi,

Alarm 5 or ES (on the display looks as E5) means that Emergency Stop button has been pressed.

To cancel the alarm you have to press twice the Cancel Alarm button (first to stop the audible and second to clear the alarm).

If the alarm keeps showing, this means that the button is stuck or the keyboard membrane has oxidation between the tracks. Either way you have to replace the keyboard membrane from the Hand Held controller.

Hope this helps.

Maia