Wed Jul 18 2018

Will not ventilate when switching from bag to vent mode.

Hello all, I have a 7100 that will not ventilate when switching from bag to vent mode. I have checked the microswitch on the lever and the GUI does show indicate which mode its in. I can here the vent engine actuating but I get no movement from the bellows and no movement of air across the flow sensors, thus I get no waveform on the GUI. I have done all leak tests in service mode and in normal operation mode, with no leaks. I have done all calibrations and diagnostic checks in service mode with no problems. The s/w is 1.3. Any help insight would be much appreciated. Thank you

   RE: Anesthesia Gas Machine Report Post
Wed Jul 18 2018Reply from Troubleshooter
avatar placemark

Please check Pneumatic connections & Settings for Drive gas.If drive gas is absent due to leak or some other Problem machine will not Ventilate.


Report Post
Wed Jul 18 2018Reply from onodayb
avatar placemark

HI Troubleshooter,

Thank you, I did confirm drive gas is set to O2 and I have 52psi at the wall. Also have confirmed pneumatic connections with no leaks present and have torn the breathing assembly apart a few times just to make sure there were no leaks there as well.


Report Post
Sun Jul 22 2018Reply from Dr. Tamim Jaber
avatar placemark
  1. EXCHANGE THE ABS WITH WORKING MACHINE
  2. DO THE SPANING CAL AND PRESSURE LIMIT SES TESTS
  3. CHECK 25PSI ON DRIVE GAS REG
  4. CHECK /EXCHANGE THE BELLOW ASSY AND CANISTER
  5. IF ALL PASS START TROUBLESHOOTING THE ELECTRONICS
  6. CONTROL BOARD 

GOOD LUCK


Dr. Tamim Jaber
Clinical engineering 
PhD Health System Administration

Report Post
Sun Jul 22 2018Reply from tcarver01
avatar placemark
I don't know if your machine has an aux CGO port, but if it does make sure it's closed.

Report Post
Sun Jul 22 2018Reply from Dr. Tamim Jaber
avatar placemark

Thanks 


Dr. Tamim Jaber
Clinical engineering 
PhD Health System Administration

Report Post
Sun Jul 22 2018Reply from Shaji
avatar placemark

Remove the drive gas check valve it gets stuck some times. Reseat it. Did you do a 2 yr pm kit replacement recently?


Report Post