acuson antaris screen problem

when I launch the device , the screen didnot blink but when use external monitor it stated. I measured the volage to main screen was55 volt is it true or there is problem with the power supply to the main screen?

   RE: Ultrasound Systems Report Post
Fri Jul 20 2018Reply from Total MedTech
avatar placemark

yeah could be,you can open it up and check the condensers 


http://totalmedtech.ca
Report Post
Fri Jul 20 2018Reply from Mike Davis
avatar placemark
Please check the fuses on the Transformer tray(F1). If this is bad the monitor will not get all 120V.

Michael Davis
Technical Support Specialist
Conquest Imaging
Report Post