Dear All System : sire mobile compact L I have a problem with this system. When I make fluoroscopy in profile's position I have the error: E2001. Please give me help to solve this problem. Best regard.

   RE: Mobile C-Arm Report Post
Mon Jul 23 2018Reply from Tomelgy
avatar placemark

Possible trouble:    Host registered a system reboot, 
    The watchdog on D1 triggered the reboot (system overload).
 
Possible help:    Faults in the system.
    Check power supply (voltage and ripple).


Report Post
Mon Jul 23 2018Reply from Kennedy Krieg
avatar placemark

MedWrench suggests contacting the following companies to help finding a solution to your problem:

RSTI

Soma


Kennedy Krieg
MedWrench - Digital Marketing Manager
Report Post
Mon Jul 23 2018Reply from semmedical
avatar placemark

Iramus:

Error E2001    System reboot after watchdog timeout:

Possible trouble:    Host registered a system reboot, 
    The watchdog on D1 triggered the reboot (system overload).
 
Possible help:    Faults in the system.
    Check power supply (voltage and ripple).

 

Manuel Casanova

semmedical@gmail.com


Report Post
Thu Oct 11 2018Reply from bilal7849
avatar placemark

Hi friends
Please help me
siemens siremobil 4 h service manual please send me
bilalmaster9@gmail.com 


Report Post
Sat Oct 13 2018Reply from Tomelgy
avatar placemark

E2001: -

Host registered a system reboot, 
    The watchdog on D1 triggered the reboot (system overload).
 
Possible help:    Faults in the system.
    Check power supply (voltage and ripple).

Check D1 PCB, connections etc.