Error code

Im getting error code 141 (not reaching temp in time) Any wisdom on finding a solution would be much appreciated.

   RE: Tissue Processor Report Post
Mon Aug 13 2018Reply from Paul Moyer
avatar placemark

Does it ever reach temp? If not it could be a defective TH2 or TH3 thermistor, also possible heater issue.


Report Post
Mon Aug 13 2018Reply from Kjb biogroup
avatar placemark

Thank you very much! I have a parts unit available if it does not reach temp and I need TH2, TH3, or whatever. Any process that you would recommend I start so I can pin point the main issue?


Report Post
Thu Aug 16 2018Reply from Ross
avatar placemark

Does it heat at all? Does the temperature of the fluid in the retort (as measured by accurate thermometer or digital probe) agree with the displayed temperature?


Report Post
Mon Aug 20 2018Reply from LarryEdgarFox
avatar placemark

Check retort heaters, front heater may be open, replace the open heater.