Injector medrad dual
RE: Contrast Media-Injector System Report Post

Hello,
I suggest checking with the below companies to see if they can assist here.
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post

The tip is likely either broken or there is a failure of the sensor inside the front tip. The replacement of the tip and the sensor is pretty involved, as there are 2 springs in the front tip which must be installed blind and under tension. I don't generally recommend this being done in the field without experience. I would recommend a depot repair. Please contact my office to discuss options in further detail. 888-667-1062
Ryan Clarke - ISS | We Are ISS | 888-667-1062
Report Post

Report Post

Take front cover off the head, put a syringe on and twist making sure the plastic flag is going into the sensor board. I had a problem where the flag was going behind the sensor board instead of into. I had to remove the plastic flag and straight it. The problem reported was the head would not always recognize The syringe being installed