error code , service manual

OPTIVANTAGE DH - ERROR 0 x 2007 AND 0 x 1007, this error occurs during the boot up sequence. And can any anyone assist me with the error code list and preferably the service manual. thank you

   RE: Contrast Media-Injector System Report Post
Fri Aug 17 2018Reply from Alex Wright
Alex Wright

Hello! I am certain that one of the below companies can assist you with this! I recommend contacting them for help:

Maull Biomedical or Injector Support and Service

 



Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com

Report Post
Fri Aug 17 2018Reply from Ryan C.
avatar placemark

1007 and 2007 indicate no commmunications between the head and console.  Suspect bad head cable, bad head ext cable, bac display cable, bad communcations or cpu boards.

Start troubleshooting by disconnecting and reconnecting each of the cable connections and check the cables for any issues/damage in general.  Next plug the head directly into the back of the power box to bypass the ext cable.

From there, further troubleshooting will require another unit to swap components between or you are looking at shotgunning parts unfortunately.


Ryan Clarke - ISS  |  We Are ISS  |  888-667-1062

Report Post
Tue Nov 08 2022Reply from RAJ MEDITECH
avatar placemark

What about 0x2042 ?


Report Post