The equipment is turn on but when begin the exam dont working and show this error, please sombody can helpme??
Tue Aug 28 2018
ERROR SER 3016
Tue Aug 28 2018Reply from Alex Wright

Hello, there is some information about errors on the service manual you can take a look at here.
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
Wed Aug 29 2018Reply from Ryan C.

Confirm that you are able to load both A and B sides without issue?
Set up an injection with B-side only (no fluid required at this time). Does the unit pass or fail?
Set up an injection with A-side only (again no fluid required at this time). Does the unit pass or fail?
Assuming it fails on only one side, you can swap the motors between sides and repeat the above test to ensure the failure is a particular motor. Repalce the motor.
Ryan Clarke - ISS | We Are ISS | 888-667-1062