Error 60,1408, no exposure using table bucky

Any help with this problem is greatly appreciated.

   RE: Analog Radiography Report Post
Fri Aug 31 2018Reply from BMET-KY
avatar placemark
error 60-1408H AEC counter error If there is no AEC feedback in AEC station exposure. 1. ION chamber 2. AEC cable, connection 3. AEC board 4. kV control

Report Post
Wed Sep 05 2018Reply from Ed Bican CRES
avatar placemark
Check your AEC cable for opens. I have had wires in the cable open due to flexing from the table moving up and down.

Report Post
Mon Oct 15 2018Reply from nathan.burks
avatar placemark

Thank you for the replies. It turned out to be a bad ion chamber.