Error 60,1408, no exposure using table bucky
Fri Aug 31 2018Reply from BMET-KY
error 60-1408H
AEC counter error
If there is no AEC feedback in
AEC station exposure.
1. ION chamber
2. AEC cable, connection
3. AEC board
4. kV control
Report Post
Wed Sep 05 2018Reply from Ed Bican CRES
Check your AEC cable for opens. I have had wires in the cable open due to flexing from the table moving up and down.
Report Post