Wed Sep 05 2018
Error 044
Hello, I have a IE33 with intermittent error 044. Tech mentioned it also said something about cooling fan. I checked all fans and they appear to free of debris and spinning correctly. Does anyone know what the error code mean its not on any error list or manual I have. Thanks
Wed Sep 05 2018Reply from rsiefers
years ago I had a similar problem, the unit would give the error code and then shutdown.
I found that the fan & temp regulator (contains an op-amp) had failed, I replaced the board and problem solved
To access, tilt back monitor and near front and down is this board
Wed Sep 05 2018Reply from dferguson
dferguson
There are 8 fans on the system if it is a UMB style system or higher.  I would look at all of them.  2 for the Acquisition Card Cage, 2 in the wheel wells, 2 for the video cards and 2 on the processors.
Wed Sep 05 2018Reply from Jim Carr
Jim Carr
If you can open the Log Browser, the ErrorLog often helps identify which fan or associated cable is causing the error.
Jim Carr
AUS
USA office: +1 918-628-2831
Sat Sep 15 2018Reply from starlin2121
Hello
Please provide me your contact info to help you
Thank you
Best Regards
Mon Oct 01 2018Reply from Set-Ing
You could attach the error log, before the error shown on the display. To identify the problem, specify the type of hardware, on the label of the equipment rules.
Specialized in US, RX