Hi.
I've been in charge of a GETINGE 533HC sterilizer ten years ago. And today suddenly, the system doesn't start and the display only shows: GETINGE PACS3000.
I tried two "Cold Start", but the device doesn´t respond.
Any idea someone?
Sat Sep 22 2018
PACS doesn't start
Wed Oct 17 2018Reply from Alex Wright
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
Thu Oct 18 2018Reply from Alex Wright
"Power down the sterilizer and reseat the cable going to the back of the display and power it back up. If that doesn’t work power down the sterilizer and reseat all cables. It is either a bad connection a bad cable or a bad control board. Usually that problem is just a bad connection."
For additional information email ReMedPar!
- Joseph Caccioppi, Service Technician at ReMedPar
Alexandra Wright MedWrench - Sales & Marketing Coordinator email me: awright@medwrench.com
Report Post
Thu Oct 18 2018Reply from CDonoso
Hi..
Thanks for your help. It seems the control board is damaged, I replaced it and is all functional now.
Greetings.