Error 107

What is Error 107? Does somebody have a service manual of the visulas 532s?

   RE: Surgical Laser Report Post
Fri Sep 28 2018Reply from ChristineMae
ChristineMae

Error 107: PowerFail. The CPU needs to be replaced, The Power Supply needs to be replaced. Or the System Control PCB needs to be replaced. Sorry the manual is very vague and gives no further instructions. If you require further assistance, you may reach out to me at info@umlasers.com. I will provide as much tech support as I am able.




Christine Ricoy

C.L.R.T.

Universal Medical Lasers