Wed Oct 03 2018

Error Code

Could you tell me what error Code DFV0008 is?

   RE: Washer/Decontaminator Report Post
Wed Oct 03 2018Reply from Brian A
avatar placemark

The DFV0008 error code is related the amount of disinfectant measured during the disinfection cycle. If the amount if disinfectant is inadequate for a high level disinfection, this error will occur. The first step is to re-run the cycle, following the procedures detailed in the user manual. For more assistance, please contact Nanosonics at 844-876-7466, opt 1.


Report Post
Thu Oct 04 2018Reply from Jritt
avatar placemark

This is a error because you mist value need calibration


Report Post
Thu Oct 04 2018Reply from Brian A
avatar placemark

Hi GGB & Jritt,

The DFV0008 error is usually associated to amount of disinfectant provided for the high level disinfection and is not normally a measurement or calibration issue. For more assistance, please contact Nanosonics at 844-876-7466, opt 1.


Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree