Mon Oct 08 2018

Fault Error 11

While we were in surgery today, we tried starting up the ellman and got an error 11 fault. I couldn't find any information in the online manual and we have reached out to the REP. Any advice?

   RE: Electrosurgical Unit (ESU) Report Post
Mon Oct 08 2018Reply from Chris S
Chris S
Error code 11 is: Foot or Fingerswitch pressed during power up protection failed. I would try a new footswitch if one is available.

Official MedWrench Guru
Report Post
Wed Oct 10 2018Reply from Clinic meyer arl
avatar placemark

Interesting, we don't actually use the foot switch. I will investigate. 


Report Post
Mon Mar 21 2022Reply from Tony Mann
avatar placemark

Hi, Does anyone know whet error code 13 is? Thank you.


Report Post
Tue Mar 22 2022Reply from T-Lobato
T-Lobato

Tony,

The service manual can be found under documents.

 

T.