What would cause the discharge levels to read high. I am getting high readings when testing on my analyzer. For example on 360J I get a reading of 373.9J.
Thank you.
Tue Oct 09 2018
Discharge levels are reading high
Tue Oct 09 2018Reply from Crule
373.9 is acceptable as the energy delivery for the LP 15. Limits are 334.9 to 384.9 per service manual. You can adjust the energy levels by following the Defib Cal in the "Service/Calibration" menu.