Airborne 750I incubator
You must first access the electronics compartment as follows:
1) Turn the incubator power button off and unplug the AC power cord.
2) Remove the inner and outer infant chambers.
3) Remove the four latches that secure the infant chamber (two screws per latch).
4) Remove the four screws that secure the lamp assembly and gently pull the assembly to expose the connector. Disconnect it and remove the lamp assembly.
5) Carefully lift the front edge of the airflow tray assembly just enough to gain sufficient access to disconnect the two electrical cables and the ground cable. Disconnect both electrical cables and the ground cable.
6) the airflow tray assembly can now be completely removed.
To remove the battery:
1) Loosen the four quarter turn fasteners securing the battery clamping bracket.
2) Remove the negative lead from the battery terminal. If necessary, the battery may be carefully lifted or tilted to allow better access to the terminal.
3) Remove the positivelead from the battery terminal. If necessary, the battery may be carefully lifted or tilted to allow better access to the terminal.
4) The battery can now be carefully removed.