F23 error on Baxter 6201 IV Pump

My associate is having trouble with an old Baxter 6201 IV pump. It gives error F23. He swapped CPU boards with no change. He was going to order a sensor board, but was wondering if anyone else had this show up before. [i]Provided by the Technation Listserv, question posted by Kevin S.[/i]

   RE: Infusion Pump Report Post
Tue Nov 23 2010Reply from MedwrenchAdmin
avatar placemark
This is an interrupt error caused mainly by the failed CPU Board. We have seen instances when the Sensor Board is to blame as well, or a combination of both. They can both be repaired and the problem can be corrected without buying a new board. [i]Provided by the Technation Listserv, answer posted by Troy G[/i]

Report Post
Tue Nov 23 2010Reply from MedwrenchAdmin
avatar placemark
Also for the F23 Error Code I have replaced the white ribbon cable between front and rear housings. [i]Provided by the Technation Listserv, answer posted by Andy A, Anaheim, CA[/i]