I wanted to share a template we use at my facility for the Vectra Genesis. Please verify all steps and make sure it fully addresses OEM standards or modify it to meet your AEM.
PM Test Results:
[X] Applicator condition.
[X] Power cord condition.
[X] Button condition and function.
[X] Ultrasound Test 1Mhz
Set point= 1.2 Meter Read back= 1.16 Spec (+/-.17W)
Set point= 2.0 Meter Read back= 1.97 Spec (+/-.28W)
Set point= 3.2 Meter Read back= 3.20 Spec (+/-.45W)
Set point= 4.0 Meter Read back= 4.03 Spec (+/-.56W)
Set point= 5.2 Meter Read back= 5.01 Spec (+/-.73W)
Set point= 6.0 Meter Read back= 5.68 Spec (+/-.84W)
Set point= 7.2 Meter Read back= 6.99 Spec (+/-1.0W)
Set point= 8.0 Meter Read back= 7.67 Spec (+/-1.12W)
Set point= 9.2 Meter Read back= 8.84 Spec (+/-1.23W)
Set point= 10.0 Meter Read back= 9.51 Spec (+/-1.40W)
[X] Ultrasound Test 3.3Mhz
Set point= 1.2 Meter Read back= 1.07 Spec (+/-.17W)
Set point= 2.0 Meter Read back= 1.84 Spec (+/-.28W)
Set point= 3.2 Meter Read back= 3.15 Spec (+/-.45W)
Set point= 4.0 Meter Read back= 4.24 Spec (+/-.56W)
Set point= 5.2 Meter Read back= 5.16 Spec (+/-.73W)
Set point= 6.0 Meter Read back= 5.91 Spec (+/-.84W)
Set point= 7.2 Meter Read back= 6.91 Spec (+/-1.0W)
Set point= 8.0 Meter Read back= 7.70 Spec (+/-1.12W)
Set point= 9.2 Meter Read back= 8.88 Spec (+/-1.23W)
Set point= 10.0 Meter Read back= 9.56 Spec (+/-1.40W)
[x] Interferential (IFC) Channel 1 and 2.
Chattanooga Box
- Set load switch to 1K.
- Plug cables into Channel 1 and 2 IN ports.
Scope
- Channel 1: 20.0V
- SEC/DIV: 250us
- Place probe on Channel 1 of box.
Unit
- Select IFC
- Vector Scan: Automatic: 40%
- Sweep: OFF
- Beat Fixed: 100 Hz
- Power/Intensity: 50
Output: Sine Wave (No Specs)
[x] PREMOD
Chattanooga Box
- Set load switch to 1K.
- Plug cables into Channel 1 and 2 IN ports.
Scope
- Channel 1: 10.0V
- SEC/DIV: 10.0ms
- Place probe on Channel 1 of box.
Unit
- Select PREMOD
- Ch. Select: 1
- Sweep: Off
- Cycle Time: Continous
- Beat Freq.: Fixed
- Beat Fixed: 100 Hz
- Power/Intensity: 50
Output: Normal/Inverted Signwave stacked(No Specs)
[x] HIVOLT
Chattanooga Box
- Set load switch to 1K.
- Plug cables into Channel 1 and 2 IN ports.
Scope
- Channel 1: 50.0V
- SEC/DIV: 25us
- Place probe on Channel 2 of box.
Unit
- Select HIVOLT
- Ch. Select: 2
- Polarity: Positive
- Cycle Time: Continuous
- Sweep: Continuous
- Frequency: 100pps
- Ramp: 0.5 sec.
- Display: Voltage
- Power/Intensity: 250
- Adjust Trigger and Horizontal positioning to get the
waveform.
Output: Positive Sawtooth Waveform(No Specs)
[X] RUSSIAN
Chattanooga Box
- Set load switch to 1K.
- Plug cables into Channel 1 and 2 IN ports.
Scope
- Channel 1: 50.0V
- SEC/DIV: 5ms
- Place probe on Channel 2 of box.
Unit
- Select RUSSIAN
- Mode: Single
- Duty Cycle: 50%
- Cycle Time: Continuous
- Burst Frequency: 100bps
- Ramp: 0.5 sec.
- CV/CC: CV
- Power/Intensity: 100
Output: Square-Flat-Square... Waveform (No Specs)