Elastography not working on linear probe
RE: Ultrasound Systems Report Post
Shear wave mode is available on the C1-6-D and 9L-D probes only, starting with R5 hardware and software.
The ML6-15-D will do elastrography. It was the first Logiq E9 probe to have strain elastography, starting at version R2 software. When you press ELASTO wth the ML6-15-D activated, there should be a STRAIN button on the touch screen to go into that mode.
I believe this will make it so you can go into Strain directly when you select the ML6-15-D and then press ELASTO: Select Utility > Imaging > General. Then select the ELASTO tab (application), and the ML6-15 from the Probe pull-down menu. Specify the Default Elasto setting to be Strain.
If that works and you like those settings, this would be a good time to do a backup of the User Defined Configuration. :-)