Hello, I'm a tech who fixes and repairs these SAM units. I pretty much know how to fix the errors but I wanna know what puts them in the state. Like, "service module ram". I know how to fix this issue but does anyone know how it gets into this state?

   RE: Medical Gas Monitor Report Post
Fri Nov 30 2018Reply from Ali Youssef
Ali Youssef

Hello, Sam Modules give you this error when you have a corrupted software. It’s a software issue on the main board of the Sam.

 


Report Post
Thu Dec 06 2018Reply from Aneserv Medical Inc.
Aneserv Medical Inc.

The microprocess watches for certain voltages that are defined in the software and when they drop below or above the tollerance you get an error. In the SRAM error, the voltage of the RAM Battery drops below the 2.7 Volts.