Statspin express 4

lid won't open and 5 min and 10 min light up with wrench. replaced rotor- still same problem- don't have manual, is it the control PCB? tried resetting rotor holding the 2 buttons while turning on--same issue. can anybody help me with this?

   RE: Centrifuge Report Post
Tue Dec 18 2018Reply from Laserman007
avatar placemark

actually the lights will blink with wrench.....if that is any more help in troubleshooting


Laserman007
Report Post
Wed Dec 19 2018Reply from MedWrenchManager
MedWrenchManager

You may find a solution in our forums for this product, here.

You can also try reaching out to Ozark Biomedical for help.


Report Post
Thu Dec 20 2018Reply from Chris S
Chris S

The red light would be steady if it was the rotor counter. 

The cause of the error condition can be identified by the combination of green cycle time indicators that are also flashing. See Error Codes Table. Looks like your error is that the rotor can not maintain speed.  Possibly the motor?

Error Codes Table

  • (3 min blinking) Failure to reach speed in 30 seconds. Confirm no obstruction in bowl area, contact customer service. 
  • (5 min blining Latch is not locked. Confirm cover is fully seated in closed position.
  • (3 & 5 min blinking)Rotor exceeded set speed. Remove and inspect rotor, reinstall.
  • (10 min blinking) Cover is not closed. Confirm cover is fully seated in closed position.
  • (3 & 10 min blinking) Out of balance. Confirm tubes are balanced in rotor and centrifuge is on a level surface.
  • (5 & 10 min blinking)Rotor cannot maintain speed. Contact customer service.
  • (All 3 blinking)Shield is not installed. Confirm shield is installed and in the locked position.
  • Continuously illuminated Error / Service indicator.  Cycle Counter 50,000 spins have been reached, contact distributor or customer

Official MedWrench Guru
Report Post
Thu Dec 20 2018Reply from Laserman007
avatar placemark

I replaced the rotor...still same problem.

control PCB?


Laserman007
Report Post
Fri Dec 21 2018Reply from C. Mash
C. Mash

The 5 and 10 minute lights designate the instrument cannot maintain speed. The most common cause for this issues is the motor drive itself. The only caveat is that the motor can stress the drive circuit and damage the main pcb. We normally start repairs with a motor and replace parts as needed. I can explain further via phone (800-457-7576). I hope this answered your questions. Thanks again for using MedWrench.


Ozark Biomedical
Senior Technical Specialist
800-457-7576
Report Post
Thu Feb 14 2019Reply from TRS
avatar placemark

First of all, does the motor spin freely?  Check for rough bearings, ability to spin freely and check for radial play of motor shaft with the unit not powered up.  If the motor is seized up, feels rough when rotated or the shaft has play, you probably need a motor.  If this isn't the case, remove the rotor and bowl, check for a broken motor wire at the terminal connections.  If a broken wire is noted, further inspection may reveal a discoloration in the connector itself or simply a broken wire.  A broken wire usually indicates that a vibration is occuring due imbalanced loads being spun, a bad motor mount or a cracked rotor/carrier.  If no connector issue is identified, closely inspect the control pcb and look for foreign debris (melted tube caps, trash) and physically damaged components on the board.  A little inspection may keep you from accessing the instrument mulitiple times and save you down time.


Report Post