ERROR 35 LORAD M-IV

Hello dears, i have an error in mammography lorad M-IV. when i try to perform exposure. system can not perform exposure and found error 35. please tell what should i do.

   RE: Mammography Report Post
Wed Jan 16 2019Reply from jay672
avatar placemark

Check and calibrate your filament waveforms, also check your filament standby voltage 


Report Post
Wed Jan 16 2019Reply from Jodiflo
Jodiflo

Be sure to check the fuses in the bottom, inside the center of the gantry (left door). 9 and 23 have been known to cause the filament standby voltage to drop to zero because they have become dislodged or the connections simply has become lost caused by corrosion or other factors. Reseat all the connections on the Filament Control board as well. You are looking for 250 mV +/-10mV as the standby voltage R74.


Jodi Andrews - Official MedWrench Guru 
Report Post
Thu Jan 17 2019Reply from Murtaza jack
avatar placemark

Hi jodiflo,

i was check voltage across the R74. (0) volt across the R74. and i was check all the fuse.

All fuse was Ok. now what should i do.


gm
Report Post
Thu Jan 17 2019Reply from jay672
avatar placemark

Replace the filament control board


Report Post
Thu Jan 24 2019Reply from Murtaza jack
avatar placemark

Hi jay how i check the filament voltage.


gm
Report Post