Changing Date and Time
RE: Vital Signs/Multi-Parameter Report Post
NOTE: In the S/5 Network, the patient monitor’s date and time is automatically synchronized
with the date and time of the iCentral it is connected to. You can set the date and time locally
only if the patient monitor is not connected to the S/5 Network and the patient monitor is in a
discharged state.
CAUTION NETWORK DEVICE TIME SYNCHRONIZATION — When adding a new device to
the CARESCAPE Network, the existing devices on the CARESCAPE Network will
synchronize to the new device’s time. To prevent potential time
synchronization issues, you should set the new device’s time to be as close as
possible to the time (within one minute or less) used by the existing GE devices
on the CARESCAPE Network.
1. Log in to Webmin.
2. Select Configuration > Time.
. In the Time Configuration window, in Configure Date and Time, update the following
fields as needed:
• Date
• Month
• Year
• Hour:Minute
• AM/PM
• 12/24 Hrs
4. In the Time Configuration window, in Configure Time Zone, select the appropriate UTC
Offset, if applicable.
NOTE: The time zone configuration only applies to communication with Citrix server.
5. Select Save.
The manual time configuration takes effect immediately.
Wendy Argueta
Edge Biomed
wendy@edgebiomed.com
Report Post
IndyWendy,
I have changed my times, in Webmin, just like you prescribed, and the time keeps slipping at a rate of -5 minutes per month. GE tech support is unknowledgeable and frustrating to deal with, so I'm seeking help outside their useless support structure. We like our field engineers, but GE phone support is THE WORST. Any insight would be helpful.
Report Post
Our technician said it sounds like the CMOS battery which should be located on the main board and may be soldered on. The battery is a watch style battery.
Bill Cahill
Bio-Medical Equipment Service Co.