Bear in mind that just because the probe fits into the connector doesn't mean it is compatible. Additionally, if they are using disposable probes will this becoma an issue. I would first ensure you have at least one "golden set" of accesories with 100% confidence that it is correct and accurate (even if this is the perfectly matched OEM set and it is expensive). Then I would turn my attention to the monitor as being the culprit.
Report Post
I would agree it sounds like you have a probe compatibility issue. The monitors can be purchased with 3 SPO2 configurations:
Generic Ohmeda SPO2, Masimo SET SPO2, and Nellcor OxiMax SPO2. You will have 1 of these options installed. I don't believe it is possible to order a monitor with no SPO2 option.
Official MedWrench Guru
Report Post
Thank you very much for the replies. You are correct. It was a probe compatibility issue
Report Post
Just because it says MASIMO on the Spo2 input side, doesnt mean it is MASIMO. Someone could have used a good case and put Nellcor in the unit. With Nellco you have two applicable pateint cables (Nellcor Oximax and Nelcor Oxismart) Look alike but are different. Possibly see if you can get a Nellcor cable & sensor.
James Vollbracht
Medical Equipment Sales Associates, Inc.
Report Post
How do you determine which brand SpO2 probe is required, if there is no label, other than trial and error, or is there some identifier in the cold boot or service menu? Any identifier inside the DAS module? Thanks