Mon Jan 28 2019

Error code CAL 2 FAIL

Zimmer - ATS 1500 showed at display Error code CAL 2 FAIL what should i do to fix this problem

   RE: Tourniquets Report Post
Mon Jan 28 2019Reply from Chris S
Chris S

Cal 2 indicates the 2nd cuff amplifier is out of calibration or Cal Valve malfunction.


Official MedWrench Guru
Report Post
Wed Feb 13 2019Reply from Chris-CBET
avatar placemark

Cal procedure is in the manual, if that doesn't work....toss it in the scrap metal and a buy a newer one, not worth the $$


Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree