Problem with network - IU22
RE: Ultrasound Systems Report Post
Good afternoon,
I try to send a 20/30 DICOM images, the sending is done to 2 DICOM servers, before the sending was done instantaneously, the network card presents miss packet when performing the ping. Any way to put another network card? If yes, how?
Report Post
Do you know what is the hardware configuration of your machine?
Did you check your network cable?
also try reinstall software first before replacing any parts and make sure you have the current system backup and take note of your network and dicom settings.
visit us @ http://www.emedicalsol.com
Report Post
Howdy,
There is not a network card in the system. The ethernet adapter is on a CPU motherboard and then is routed through the AVIO board in the rear of the system.
Check the network settings on the system. You might also try connecting the system network output to a hub or switch first, then connecting that switch to your network. There was a design problem with some iU22/ iE33 systems where the ground on the ethernet output would not be at 0V, which caused network connectivity problems.
Good luck,
Jim Carr
Director of Service & International Operations
AUS
Toll Free (USA, Puerto Rico and Canada): 866-620-2831
USA office: +1 918-628-2831
Email: Service@advancedultrasound.com and Sales@advancedultrasound.com
Report Post
Hello, Really is the card that is with problem, I added a PCI network card and I did the ping test and it worked correctly, would have how to put that PCI card to work in the device software inves of the card that is in problem? Any configuration that I have to do within the Software for it enexergar this PCI card as main?