Will not draw a vacuum on the chamber
RE: Sterilizers/Autoclaves Report Post
Dear Matt D.,
We came across your post about the AMSCO 400 Series Small Steam Sterilizer. We have a dedicated Service Support team who can help. Please provide us with your preferred method of contact or call 1-800-333-8828 for assistance.
Sincerely,
STERIS Corporation
Report Post
Assuming you replaced all check valves, including the 2 on the heat exchanger and the 1/4" NPT swing check coming from the exhaust manifold, rebuilt all the valves and checks in the 3 manifolds, replaced the ejector, and rebuilt both traps, (mainly the jacket trap, including the seat) and verified your transducer, calibration, and control, then my next check would be the door gasket leaking steam into the chamber during vac. Bad gasket, dirty channel, or both.
What are your vacuum levels and what is the temp reading when it bottoms out? Does a lot of steam come from the chamber after opening the door upon failure?
Oh, you did check the chamber drain strainer inside the chamber, right?
Report Post

Sorry for not letting you guys know. This items is repaired. S1 was intermittantly sealing. Coil was good but the diaphragm assembly was bad.
Matt