Will not draw a vacuum on the chamber

Initial issue "Too long to evacuate". Water pressure 35PSI, replaced all components kits , seats, coils, injector, heat exchanger, check valves. verified pressure transducer is working with in specs. Steris is confused can't find the problem. WTF, any ideas?

   RE: Sterilizers/Autoclaves Report Post
Mon Feb 11 2019Reply from Kara Crissman
avatar placemark

Dear Matt D.,

We came across your post about the AMSCO 400 Series Small Steam Sterilizer. We have a dedicated Service Support team who can help. Please provide us with your preferred method of contact or call 1-800-333-8828 for assistance.

Sincerely,

STERIS Corporation


Report Post
Wed May 15 2019Reply from Chris35
avatar placemark

Assuming you replaced all check valves, including the 2 on the heat exchanger and the 1/4" NPT swing check coming from the exhaust manifold, rebuilt all the valves and checks in the 3 manifolds, replaced the ejector, and rebuilt both traps, (mainly the jacket trap, including the seat) and verified your transducer, calibration, and control, then my next check would be the door gasket leaking steam into the chamber during vac. Bad gasket, dirty channel, or both.

What are your vacuum levels and what is the temp reading when it bottoms out? Does a lot of steam come from the chamber after opening the door upon failure?

Oh, you did check the chamber drain strainer inside the chamber, right?


Report Post
Mon May 20 2019Reply from Cbet_Matt
Cbet_Matt

Sorry for not letting you guys know. This items is repaired. S1 was intermittantly sealing. Coil was good but the diaphragm assembly was bad.



Matt

 
Report Post
Mon Feb 17 2025Reply from Chris-CBET
Chris-CBET

your water pressure is too low btw. should be at least 40 dynamic