Norm to Mag 1,2 issue
RE: Mobile C-Arm Report Post
It could be a corrupted cal file. I would go through the collimator cal in Utility Suite. The collimator iris blades are set here. Remember to save the file to the system and then reboot and test it. Good Luck
Art
Report Post
p.s. i forgot. change the coin battery on the generator interface board. this is where the calibration data is stored.
Art
Report Post
if its not the cal file, the look at the signals from the FFB to the Collimator/II especially throught eh HV cable. Yeah change out the coin battery should be greater than 3VDC
Report Post
If you measure the image size & compare with correct size is the Mag1 size wrong, if yes it could be the II power supply.
If you are referring to the collimato Iris smaller but image size correct, then could ba an adjustment cable etc. If you look at the Iris collimator can you open and/or close without problems!
Tom
Report Post
If you have not already figured this out it sounds like the wire that goes through the High Voltage control cable that selects the desired II size is broken. If neither II size works I think there may also be a ground wire associated with both of them. Look at the C-arm interconnect schematic. Ohm out the White/Orange and White/Yellow wires between P7 on the II power supply and P8 on the backplane PCB. If one or both of them are broken there are unused wires in that cable you can splice in to repair it without having to replace the entire HV control cable assembly.
Report Post
There are spare wires, i am not sure which one but let me look up my notes and get back to you this morning.
Report Post