Tissue tracking device ASP 300 S finishes the process smoothly.
But I start STANDART cleaning program ( Bottle 11: Cleaning Xylene cycle 11 and 65 degree , Bottle 12: Cleaning Alcohol 6 Cycle 50 degree, Bottle 13: Cleaning Water-Alcohol 5 Cycle 50 degree) .
ASP 300 S warning fault step 12 only. Error Mesaage is 452 : The Software has detected overpressure during an operation. Switch the power off and back on to clear this condition and contact your service organization before using the ASP 300 again.
Then the system resets itself. Cleaning program resumes and completes.
PLEASE HELP ME ! What is problem?
Sat Mar 30 2019
I get a warning message
Sun Mar 31 2019Reply from elipow
It’s most likely not over-pressure. That alarm actually trips in event of over-pressure OR over-temp. Your time of failure is at the hottest that instrument ever gets.
Report Post
Sun Mar 31 2019Reply from trinhbk04
Mon Apr 01 2019Reply from umitedi
Elipow Thank you for your answer. Yes true. I think this problem so I find solution . Give me mail adres please. I will write you