help needed on CR 85-X Type 5148/100

Error : 24AB0 & 20E0E

   RE: Computed Radiography Report Post
Mon Apr 08 2019Reply from arnaudboub
avatar placemark

24AB0
E_HDC_IPVACPO_TAKEIP
Meaning : Handling Control Module:  Vacuum error postscan.
Reason    : Got no vacuum  while taking IP from scanner.
          Probably IP Jam.
Cure    : Open doors and put IP into cassette.

20E0E
E_SCN_FN_TIV_D_RANGE
Meaning : Pixel values with "dark" photomultiplier are out of range.
Reason    : Too much background light or photomultiplier box defective.
Cure    : Avoid too much background light, keep door closed at start.
      If problem persists, replace photomultiplier box.
 

 