Thu May 23 2019

Operating parameters

Does anyone know the operating minimum/maximum temperature and humidity for the C.A.T.S ?

   RE: Autotransfusion Unit Report Post
Wed Jun 19 2019Reply from Ray Brown, CBET
avatar placemark

I had a manual for the older CATS somewhere, but checking with the specs on the new CATS,  your temperature should be in the range of +15C to +27C, and humidity from 30% to 75%, with a temporary allowance up to 95% humidity.

The unit can be stored at a range of -15C to +50C.

This is for the Fresenius Kabi CATSmart, but it's essentially the same mechanism with a different computer, so these values should be the same.