Tube is hot even on idle
RE: Mobile C-Arm Report Post
Are they doing alot of HLF or cine runs that will over heat a tube fast. Draping are they draping in a way that the fans are not getting enough air, in or out. Enviorment of the room do they keep it warmer than usual, seen where a doc likes it at 80 degrees. Fan filters clean?
Report Post
This is an outpatient surgery center so they really use this carm alot. Im not sure about HLF or Cine, but Im sure they use those features. I will check on the draping. The fan does work and the filter is clean. The room is feeling a bit warmer than usual. Thank you sir! I need to go back on site and evaluate the tempeture in the room.
Report Post
Sounds like the stater / filament is always on when it should only be active during actual use. The filament needs to pre-heat / charge before the tube load is exercised and an exposure is made. If always on it may generate excessive heat and shorten tube life.
Report Post
Hello I suggest you redo all tube calibration. Including the preheating of the filament.
Report Post