Error code

Hi. I have a Problem with the 5012 error, if someone could help me or send me the Service manual, thanks!

   RE: Mobile C-Arm Report Post
Tue Jun 04 2019Reply from Tomelgy
avatar placemark

Error 5012 is: The maximum kV values were exceeded.

Could be X-Ray tube (filament)

Could be X-Ray tube (arcing)

Could be cables to X-Ray tube

Could be D2 PCB failed

Measure kV + & kV -, should be symetrical

Does the error appear at high, low kV's?

Select: kV stop & drive kV to minimum, are you able to radiate without error? If no error appears at fixed low kV, slowly increase kV to see at what point it fails, if it fails around 50 to 60kV it could be one leg of the feedback transformer (in X-Ray tube/Sirephos failed).

You could also monitor the kV+ & kV- test points during tests.

Tom

 

 


Report Post
Tue Jun 04 2019Reply from Lessliefl
avatar placemark

Thanks for answer me.

I already did that Tests, but, if fails in 60kv , and showed up the 4028 error, do you know if these errors are conected? 

And you can show me in which Part of the Equipment the transformer is? 

Do you have Any way yo contact you? 

Email or phone


Report Post
Wed Jun 05 2019Reply from Tomelgy
avatar placemark

Error 4028 is comming from the Memoskop (image system) as a result of error 5012

All the transformers are inside the Sirephos (X-Ray tube monoblock).

At this stage it could be D2 PCB, Cables or X-Ray tube. Do you have another tube to try/test? You can place a loan X-Ray tube next to existing tube and transfer power cable/connector.

My experience is X-Ray tube, but check before replacing such an expensive part.

What is your email?

Tom


Report Post
Wed Jun 05 2019Reply from Lessliefl
Mon Nov 25 2019Reply from Erin Register
Erin Register

Although it is your choice to go offline with your discussion, we are sure that other MedWrench users would appreciate if you kept your discussion within the thread. This will help to keep a permanent record of the issue you are facing and how it was solved to help others in the community that may be facing a similar challenge.  


Report Post
Mon Nov 25 2019Reply from outinleft
avatar placemark

Error F5012 Gen. kV max exceeded

Possible trouble: The maximum KV values were exceeded.

Possible help: KV actual value acquisition defective (board D1).

Possibly board D2 defective.

Oscilloscope + KV and - KV. Values must be symmetrical.

(Measuring divider: 10 KV corresponds to 1 V). 

With asymmetric KV values check actual value cables, possibly Sirephos defective. 

 

 


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Wed Jun 22 2022Reply from Fahim4221
avatar placemark

I have error 4001 on siromobile 2000, anyone know the error ? Thanks an advance 


Report Post
Wed Jun 22 2022Reply from Tomelgy
avatar placemark

Error 4001 is console communication error.

FEPROM checksum. Wrong data from console to Host.

Could be a console failure, the suggestion is try downloading console software & see what happens.

Do a backup first.

Tom