I have a unit that is dropping intensity from 5 to 1 within seconds after selecting. Any suggestions?
Thu Jun 06 2019
Intensity Drop
Thu Jun 06 2019Reply from Priebe
check the touch pad if the bubble for intesity 1 is flat it is like someone pushing on it if you have a small suction cup you can try pulling it back and see if the intensity stays where you selected it. If you have two light swap the connection and see if the problem follows to the other light head
Thu Jun 13 2019Reply from Chris-CBET
check for a stuck key, failing power supply. Has thee been any liquid, humidty issues before this happened? Any hardware changes?