I Need service manual of envision ct

   RE: Contrast Media-Injector System Report Post
Mon Jun 10 2019Reply from Ryan C.
avatar placemark

I don't have a service manual I can offer, But the error itself points to a suspect issue with the potentiometer.  Advance the ram 1" and cycle power to see if ti clears, run operational checkout.  Check and reseat all cable connections.  Calibration and/or replacement of the potentiometer.  Repalcement of the head control card. 


Ryan Clarke - ISS  |  We Are ISS  |  888-667-1062