operations manual
RE: Centrifuge Report Post
The S2210 has not been manufactured since 2012. The operators manuals for the later versions (2010 and later) are still available.
Report Post
Will not start when the start switch is pressed and “HELP
1.1” is displayed in the Speed and Time Displays.
POSSIBLE CAUSES POSSIBLE SOLUTIONS
Indicates no power is getting to
motor.
Ensure the motor wires are properly
connected to MOTA1, MOTB1, and
MOTC1 on the Power PCB.
The brake/drive relay’s drive contacts
may be pitted or worn.
SYMPTOM
The motor tachometer cable is
unplugged.
Ensure the motor tachometer cable is
connected to MOTOR1 on the Power
PCB.
Report Post
Will not start when the start switch is pressed and “HELP
1.1” is displayed in the Speed and Time Displays.
POSSIBLE CAUSES POSSIBLE SOLUTIONS
Indicates no power is getting to
motor.
Ensure the motor wires are properly
connected to MOTA1, MOTB1, and
MOTC1 on the Power PCB.
The brake/drive relay’s drive contacts
may be pitted or worn.
SYMPTOM
The motor tachometer cable is
unplugged.
Ensure the motor tachometer cable is
connected to MOTOR1 on the Power
PCB.