OEP pump trouble shooting

Hi, Does anyone have information on the Olympus flushing pump. I have one that is not delivering enough fluid. No kinks or leaks or blockages.

   RE: Medical Printers/Imagers Report Post
Thu Jun 20 2019Reply from Cipher
Cipher

Can you verify the exact flushing pump model you have?   The OEP series posted here is for a printer.   The pump should be OPF series.   The flushing pump head assembly has a finite life, typcially 2 to 5 years.   There's a performce test you can run to verify output and determine if the pump head is worn or not. The good news is that the pump head is a replacable assembly and can be done quite easily in the field without tools.  I have instructions for both the output test and the head replacmeent, but I would need the model number your have to verify if its applicable before sedning it to the MedWrench Admin for posting.   -  Cipher